The one on the top is a 3 wire RTD. The one on the bottom, I believe, is a dual element RTD. That one is a little odd to me, because of the marker colors. Usually a dual element 6 wire RTD is really only two three wire RTDs. If I had to guess, the two elements are side-by-side, and the red wire is the "signal" wire and the green on either side is the common. They are connected together at the RTD itself.
If you google "difference between 3 wire and 4 wire RTD" you'll find lots of info about how lead length compensation works. If you have a good multimeter, you can figure out the wiring it's using. Measure the resistance of the two green wires on either side of the red one. They should read zero. Then you should see the 100 ohm platinum resistor from red to either green. You should see no connection at all between each of the two RTDs. That is, two green wires will be "shorted", the other two will be as well, but infinite resistance between the two groups.
For process measurement, you really only need a 3 wire RTD. 4 wire RTDs offer better compensation, but that's usually only done in the lab.
Regards,
Mike Sharp
-
3-Wire vs. 4-Wire RTDs
Not a "dual element". Usually in a process setting like a brewery, 3-wire RTDs are considered "good enough". As stated above, 4-wire is for extreme precision and you don't need that. You can find lots of information about how to use 4-wire in a 3-wire circuit from your friends at Google. It's not difficult. Good luck.Phillip Kelm--Palau Brewing Company Manager--
Comment
-
Originally posted by gitchegumee View PostNot a "dual element". Usually in a process setting like a brewery, 3-wire RTDs are considered "good enough". As stated above, 4-wire is for extreme precision and you don't need that. You can find lots of information about how to use 4-wire in a 3-wire circuit from your friends at Google. It's not difficult. Good luck.
It can be verified with a multimeter, as I described above.
Regards,
Mike Sharp
Comment
-
You are indeed correct.
I've missed the six wires--my mistake. Looked in the first picture like three wires with a jumper to field wiring. In this case, throw it away! A new 3-wire RTD isn't expensive. I've never toyed with a 6-wire, or "dual element" RTD. I wouldn't want to unless someone tells me that they are either less expensive or easier to wire. That isn't likely. We don't need extreme temperature control, and 6 wires have to be more difficult to connect, so stick with 3-wire. By far better than 2-wire, almost same price, and standard in almost all similar process controls. Or figure out how to use one circuit of the two? Dunno. Sorry for the mistake. Be happy for someone to tell me what this is and why/when it would be used.Phillip Kelm--Palau Brewing Company Manager--
Comment
-
Both of those are 3- wire RTDs--the lower one just has two RTDs in one probe/housing. You can easily use just one of them, but they might be at different depths within the probe, so find out if so, and use the one that's deepest.Timm Turrentine
Brewerywright,
Terminal Gravity Brewing,
Enterprise. Oregon.
Comment
-
If they came from the existing equipment in the background, the dual element RTD may be driving two different signals. E.g. a chart recorder and a process control signal. I've also seen dual RTDs for redundancy, but that's more likely in more critical process control like a pulp mill. But, if they came from existing equipment, the single and the dual may not be interchangable. Hard to know without more info from the original poster.
Regards,
Mike Sharp
Comment
-
What?
Is that why someone would use a "dual element" RTD? A chart recorder or some other device? I've always used a single RTD and a controller to process the signal and drive other devices--like a recorder--from a 4-20 mA output. Or from a retransmit set of terminals on the controller. So I guess I've learned something today. Thanks!Phillip Kelm--Palau Brewing Company Manager--
Comment
-
Originally posted by gitchegumee View PostIs that why someone would use a "dual element" RTD? A chart recorder or some other device? I've always used a single RTD and a controller to process the signal and drive other devices--like a recorder--from a 4-20 mA output. Or from a retransmit set of terminals on the controller. So I guess I've learned something today. Thanks!
Given that breweries probably don't need extreme accuracy like you'd find in a standards lab, it's kind of surprising this type of RTD would be used. But again, I don't know the context of the OP's question. Are they RTDs he picked up somewhere, and wanted to use on his brewhouse? Or did they come off the brewhouse, and he needs to replace one?
Regards,
Mike Sharp
Comment
-
Originally posted by rdcpro View PostGiven that breweries probably don't need extreme accuracy like you'd find in a standards lab, it's kind of surprising this type of RTD would be used. But again, I don't know the context of the OP's question. Are they RTDs he picked up somewhere, and wanted to use on his brewhouse? Or did they come off the brewhouse, and he needs to replace one?