Hey everyone,
We just started playing with a luminometer. Today I soaked a PRV in ~1,000 ppm peracetic acid for 40 minutes. The meter read 58 RLUs when I swabbed it. Soaked for another 15 minutes - 110 RLU. Another 15 minutes - 78 RLU. The manufacturer recommends treating anything over 30 RLU as a fail. For those of you using a luminometer, do you have standards?
Thanks.
We just started playing with a luminometer. Today I soaked a PRV in ~1,000 ppm peracetic acid for 40 minutes. The meter read 58 RLUs when I swabbed it. Soaked for another 15 minutes - 110 RLU. Another 15 minutes - 78 RLU. The manufacturer recommends treating anything over 30 RLU as a fail. For those of you using a luminometer, do you have standards?
Thanks.